Google
 

PDA

View Full Interactive Version Of This Page : Connecting Concentrators


getdul981
2009-02-26, 12:58pm
I have two oxygen concentratorsto use. One is a 10 lpm and the other is a 5lpm. Both are retired medical devices that I got for a REALLY good price. (free) I have tried connecting both together, but when I do, the 5 lpm drops down to zero. Does anyone know of any way I can get the maximum benefit?

Greg

ArtcoInc
2009-02-26, 1:23pm
In addition to the flowrate (lpm), you need to know the pressure output (psi) for each machine. If one machine has a higher pressure output than the other, it will overpower the weaker machine (and possibly damage it).

If the two machines have significantly different output pressures, then I'm afraid that you will not be able to use them together.

Malcolm

NMLinda
2009-02-26, 1:54pm
This may be a dumb question, but is there any utility in putting reverse-flow check valves in line with either or both concentrators to avoid having the higher capacity unit from overpowering the lower unit? Or would that restrict the output of both to the point where it isn't worth bothering?

Linda

ArtcoInc
2009-02-26, 2:28pm
This may be a dumb question, but is there any utility in putting reverse-flow check valves in line with either or both concentrators to avoid having the higher capacity unit from overpowering the lower unit? Or would that restrict the output of both to the point where it isn't worth bothering?

Linda

The reverse-flow check valve will prevent the stronger machine from damaging the weaker machine. But, the weaker machine will still not have enough pressure to be able to contribute to the total output.

Look at it this way .... Let's say that the stronger machine puts out 10 psi, and the weaker machine puts out 5 psi. Both machines are T'd together. The hose between the two machines is pressurized to 10 psi (from the stronger machine). The weaker machine does not have the pressure (at 5 psi) to push against the pressure already in the hose (that is pushing back at 10 psi).

Does this make sense?

Malcolm

Dale M.
2009-02-26, 6:45pm
Its like "static pressure" in ventilation , "force applied" (weak concentrator) can not over come "resistance to flow" (high pressure from strong concentrator), there for flow of weak concentrator will be reduced to its ability to work against the "static pressure"...

Concentrators need to be as close to an equal match in pressure as possible to get maximum volume..

Dale

NMLinda
2009-02-26, 6:50pm
Malcom, Dale, thank you both for responding, even though I basically asked the same question Rymnoceros asked recently. Yup, what you both said makes sense. Doesn't matter if there are reverse-flow check-valves or not, same unequal pressure problem regardless.

Linda

getdul981
2009-02-26, 9:57pm
Thanks everyone. I had a feeling that the larger was overpowering the smaller and there wasn't anything I could do about it. Sounds like that's the concensus.

Greg